5
$\begingroup$

I am trying to design a high pass filter to remove baseline drift from an ECG signal. the baseline drift is of very low frequency like 0.3Hz or so with an amplitude of 25% of the ECG signal. But the bandwidth of ECG signal itself is 0.5Hz to 150Hz. I have been trying fdatoolbox in matlab to design the HPF but I was not able to remove the baseline drift accurately.Please help on how can i achieve a low order filter to eliminate baseline drift. I have got it using higher order filters of order above 1000. But I'm looking for a way if I can get lower order filter,like order of less than 30, to do the same.Thanks for your help..

$\endgroup$
2
  • $\begingroup$ What's your sampling frequency? $\endgroup$
    – Matt L.
    May 21, 2014 at 7:58
  • $\begingroup$ @MattL. you mean the sampling frequency that i've given during filter design? I've given it as some 350Hz like that.. $\endgroup$
    – Harsha
    May 21, 2014 at 8:53

2 Answers 2

6
$\begingroup$

You usually use a DC notch filter. This is a recursive filter that should kill a very narrow band of frequencies around DC. One possible implementation is a first order DC notch with transfer function

$$\frac{1 - z^{-1}}{1 - \lambda z^{-1}},$$

where $\lambda$ is some number very close to 1, say 0.99. A to demonstrate this in MATLAB,

b = [1 -1];
a = [1 -0.99];
freqz(b,a);

enter image description here

Notice that the phase response around DC gets pretty non-linear. Such is the nature of these recursive filters.

The closer you make $\lambda$ to 1, the narrower the notch gets and the more distorted your phase response becomes around DC.

For ECG signals, this will usually do. Experiment with $\lambda$ values that will suit your needs. Sometimes, if you make the notch too narrow, you won't get all off the drift, so 0.99 is usually a good starting point.

$\endgroup$
1
  • $\begingroup$ what other implementations of the DC notch exist? anything with better linearity of phase? $\endgroup$
    – Mark
    May 22, 2023 at 19:41
5
$\begingroup$

Recently, we have released a baseline filtering and noise removal method that takes into account the sparsity (and the potential asymmetry) of one-dimensional signals. It was initially applied to analytical chemistry data (chromatograms), but could possibly be used of ECG. It relies on a recursive low-pass filter, an asymmetry parameter, and a penalization related to the noise level and the strength of the signal and its derivatives.

One illustration for BEADS background removal is: enter image description here

The code for BEADS (Baseline Estimation And Denoising with Sparsity) is available in Matlab, R, C++. It has been used for instance in the following papers:

$\endgroup$
5
  • $\begingroup$ Is this approach limited to cases when chromatograms begin and end with 0 intensity? I have issues when it's not the case: github.com/tvijverb/BEADS_Baseline_correction/issues/1 But maybe I just didn't set it up properly.. $\endgroup$ Oct 29, 2018 at 15:27
  • $\begingroup$ In this implementation, you are correct, due to initializations for the IIR filter. A little preprocessing of the data helps in this cases. What do your data look like at the edges? $\endgroup$ Oct 29, 2018 at 15:31
  • $\begingroup$ There is a screenshot attached if you follow the link - it starts with non-zero baseline (presumably - a residual from previous runs) and ends with non-zero baseline. By re-processing - do you mean adding artificial 0's at the beginning and at the end? $\endgroup$ Oct 30, 2018 at 7:48
  • $\begingroup$ I do not have a full answer to all cases, honestly. Here, the jump at $0.25$ represents an issue with the algorithm hypotheses. Does it work better on the $[0.25 \, 2.00]$ interval? I am working on different preprocessing techniques (mirroring, linear slope removal), I can include your data in tests $\endgroup$ Oct 30, 2018 at 8:48
  • 1
    $\begingroup$ It works great from 0.5 to 1.25 though the baseline doesn't touch 0. $\endgroup$ Oct 30, 2018 at 9:46

Your Answer

By clicking “Post Your Answer”, you agree to our terms of service and acknowledge you have read our privacy policy.

Not the answer you're looking for? Browse other questions tagged or ask your own question.